No, everything is fine with your device. It needs to be reflashed, most likely through a programmer. Or from Ubuntu. I had the same situation, you can read in this thread: |
Beta Was this translation helpful? Give feedback.
-
So I was able to get it programmed and had to use the ST-link method unfortunately, but im not sure if I didnt set it right and my counter is all over the place it never dips below 0.500 uSv/h and often goes way up to over 1.000uSv/h which is not possible there is nothing that would give off that much everywhere I go, is there a way to tell if my tube is bad or something |
Beta Was this translation helpful? Give feedback.
-
Your HV might be too high. I would suggest you go to a custom HV profile in your Geiger settings and start with a real low value, like 5 kHz and 1% PWM duty cycle. Then check if your counter readings stabilize at normal values. |
Beta Was this translation helpful? Give feedback.
-
Thank you for the help, so I tried 5kHz and 1% but it just did nothing after that no measurements, I went back to the standard and it started working again, its weird the meter is always measuring high and sporadic no matter where I am at or what im doing, sometimes its lower at 0.300 and randomly jumps up to 1.300 then back down to 0.600 its just all over the place, Ive tried messing with background compensation which does not make much difference. It just seems horribly inaccurate and im very doubtful its Radpro as it was very squirrely before I switched over to this much much better firmware, I was hoping this would make things more accurate and stable but im not sure what else to assume other then the tube is just shitty or something else is causing erroneous readings, its using a M4101 tube and set to such in the geiger settings |
Beta Was this translation helpful? Give feedback.
-
No offense, but have you double checked correct polarity of your geiger tube? (I had that problem once) |
Beta Was this translation helpful? Give feedback.
-
Converting this issue to a discussion (#32). |
Beta Was this translation helpful? Give feedback.
-
A voltage scan can easily be done with my software GeigerLog. The picture shows a recent scan with a FNIRSI GC-01 (yellow) and firmware radpro 2.0.2. The computer does modify the voltage automatically. When the voltage goes up and down between 300 and 800 V, you can see how the countrate creeps up and down the Geiger plateau (the "plateau" is never perfectly flat). Download GeigerLog pre-release 1.5.1pre15 from https://sourceforge.net/p/geigerlog/discussion/devel2/ |
Beta Was this translation helpful? Give feedback.
-
@alfmck : if you are using pre15 as downloaded, it should already be configured for automatically using Anode Voltage Control! You will get infos printed into NotePad when logging. |
Beta Was this translation helpful? Give feedback.
-
I have now a blue-case GC-01 with a Geehy APM32F103RBT6 chip, and oddly enough, it behaves differently from my yellow-case GC-01 with a CH32F103R8T6 chip from WCH. The Geehy unit performs more poorly than the device with WCH chip! I installed RadPro 2.0.2 on both devices; easy job on both. And judging from the interaction of GeigerLog with either device, the firmware acts the same way on both. In particular, when I use GeigerLog via RadPro to set PWM (Freq, Duty), I can observe at PA3 of both chips that the PWM signal is indeed what I attempted to set it to at within 1% or better by using an oscilloscope. Comparing both circuit boards visually, I do not see any difference! Can someone else confirm or point out differences? However, the Geehy board performs VERY differently with respect to anode voltage, and I don't like it. Next pic shows a scan where GeigerLog used the RadPro firmware to step through the PWM frequency (red curve) at a duty of 50%, reads the counts reported by RadPro (approx CPS=40, orange curve), and also ready the anode voltage using an OW18E DMM via Bluetooth (green curve). Unsurprisingly, no counts are created when the anode voltage is too low. The max anode voltage reached is 600V at 16 kHz and 50% duty (magenta dotted line). Increasing freq further leads to voltage decrease, with an odd saw-tooth pattern, and finally a collapse at near 70kHz. Looking at the full curve, the factory default (490V) PWM setting is at an odd place (56.7kHz, 50%). From my perpective a workable range is black-circled allowing 350...600V (4...16kHz, 50%). This covers voltages for many tubes, but it is far away from what can be done with a counter with WCH chip, namely 1200 V !!! . Lateron I show that a freq from 100 to 600 Hz, can cover an anode voltage range from 300V to 800V smoothly. So, for setting 490V we have the options for PWM settings:
The settings are grossly different despite: Firmware is the same version, PWM signal from the chip is the same, Circuit board (looks like | is) the same. How comes? For completeness I did a quick scan of the duty from 0 to 100% at a freq=15 kHz; result shown in next pic. Max voltage reached remains at 600V. |
Beta Was this translation helpful? Give feedback.
-
Is somebody able to explain what this R38 does? In my 2 devices I found these implementations:
No difference. This cannot be the cause for the big difference in max voltage. |
Beta Was this translation helpful? Give feedback.
-
Good to see it is working! However, summarizing the 3 available control curves for a GC-01, we now got this: Strange, to say the least. And then there are all these other devices which do run RadPro ... The curves for GC-01 devices WCH and GEEHY were generated with Duty=50%. Yours were done with 2.25%. Can you do a Freq scan also with 50%? |
Beta Was this translation helpful? Give feedback.
-
I have added your new data to the graph. The new high-frequency curve (brown) is too flat for being a good control curve; the low freq curve (green) is ok. The difference between blue, red, green is amazing - it will be hard to define some generic formula! I'm beginning to wonder whether it has anything to do with the RadPro version .... |
Beta Was this translation helpful? Give feedback.
-
@ugopapetti : Now that you said it ... Just a minute ago I posted on the difference between these two GC-01 versions here: A third order parabola does an excellent fit for the Freq vs Anode Voltage function, but a much simpler linear approximations for the two devices suffices:
Makes for some frustration in programming. Is there any electrical reason for preferring one over the other? |
Beta Was this translation helpful? Give feedback.
-
Does somebody know, how HV circuits are constructed in FS-5000? According to https://github.com/Gissio/radpro/blob/main/docs/devices/Bosean%20FS-5000/peripherals.md |
Beta Was this translation helpful? Give feedback.
-
I have tired connecting The GC-01 to my computer and when I plug it in I get a windows message stating there is an issue with the USB device, tried with several cables, I opened Device manager and when I hold power down it does seem to have something change on Device manager as it resets but im just not sure what it is. Do I need to have some specific driver or is there something wrong with my Geiger counter?
Beta Was this translation helpful? Give feedback.