some things to consider: have you considered purchasing a test sample source to make a more accurate check? these can be purchased from ebay or salvaged from old smoke detectors. or maybe locally you have an area with high background you can use to check? old glassware, clocks and stone can give readings. then see which responds properly. |
Beta Was this translation helpful? Give feedback.
-
There are ways you can measure the tube voltage using a multimeter and a resistor, and they are shown on the forum (Discussions). From what I understood one meter is using M4011 tube, and the other one J321 tube. Are you sure about the tube types? Are they correctly configured in software? Finally, 0.1 +/-50% and 0.2 +/-50% are not that far from each other. One says that the radiation level is between 0.05 and 0.15 uSv/h, while the second says it is between 0.1 and 0.3 uSv/h... |
Beta Was this translation helpful? Give feedback.
-
don't forget that the readings of 0.1 and 0.2 uSv are tiny and within the error limits of these devices - if it were 200 and 600 i'd be concerned. it's worth investigating further but don't get hung up on the "3 times the reading" - different tubes can have different baseline readings. find an actual source and you'll see that it's probably not a huge difference. WARNING: |
Beta Was this translation helpful? Give feedback.
-
makes sense. |
Beta Was this translation helpful? Give feedback.
-
To compare two tubes, use CPS or CPM units. Also, use the averaging mode and get an average over 30 minutes or more. This will let you do a much more accurate comparison. If the difference is below 50%, you are fine, as the M4011 and J321 tubes are quite similar. At first sight, however, the difference seems to be larger, most likely due to a problem with the high voltage generator in the device with the higher reading (background levels of radiation are usually 0.1-0.15 uSv/h). In this case, you need to set up a custom HV profile by following these instructions: https://github.com/Gissio/radpro/blob/main/docs/users.md#hv-profiles As other users mentioned, be careful when working with high voltage, as it can be lethal. |
Beta Was this translation helpful? Give feedback.
-
Beta Was this translation helpful? Give feedback.
-
Friends, hello everyone. Another Fnirsi GC-01 radiation detector has just arrived to me. The previous one helped me restore Gissio, but before that I thought that it was damaged forever, so I ordered another one.
The question is what. Why do devices of the same models display different data (seen in the video). On the left is a new device, based on an APM processor and with an M4011 handset. On the right, my old device, with the CH32F processor. The difference in readings is almost three times. Why is this happening? How to fix this? Which of these devices is lying?
video_2024-09-18_12-18-54.mp4
Beta Was this translation helpful? Give feedback.